ICU · equipment-physics
ICU Electrical Safety — Comprehensive (Macroshock, Microshock, Equipotential Bonding)
Also known as Electrical safety · Macroshock · Microshock · Equipotential bonding · Isolated power supply · Leakage current · Cardiac floating equipment · Type CF equipment · Diathermy safety
ICU electrical safety — the principles and precautions to prevent electrical injury to patients and staff. Two types of electrical hazard: MACROSHOCK (external current passes through intact skin → body acts as volume conductor → current spreads → VF if 100 mA reaches the heart — prevented by grounding, RCDs, double insulation) and MICROSHOCK (tiny current ~10-100 μA applied DIRECTLY to the heart via a conductive pathway — central line, pacing wire, saline column → VF at 1000x lower threshold than macroshock — the KEY danger in ICU and cardiac areas). Prevention: EQUIPOTENTIAL BONDING (all conductive surfaces connected to same ground potential → no voltage difference → no current through patient), ISOLATED POWER SUPPLY (IPS — transformer isolates supply from ground → no path for leakage current → eliminates macroshock), EQUIPMENT CLASSIFICATION (Type CF — cardiac floating — leakage <10 μA — safe for direct cardiac connection; Type BF — body floating — safe for external use; Type B — basic — NOT cardiac-safe). Leakage current monitoring: all ICU equipment must be tested annually for leakage (<500 μA for chassis leakage, <10 μA for patient leads in CF equipment).
Practise this topic
On this page
Study tools
Your progress
Saved on this device.
Target exams
Red flags
- ANY conductive pathway DIRECTLY to the heart (central line, pacing wire, PA catheter) = MICROSHOCK RISK — the heart requires only 10 μA (0.01 mA) to develop VF — 10,000x less than the macroshock threshold
- NEVER use non-CF classified equipment with cardiac-connected patients — only Type CF (leakage <10 μA) is safe for direct cardiac connection
- Diathermy near pacemakers can cause electromagnetic interference → pacemaker inhibition or reversion to fixed-rate mode → use bipolar diathermy (current path confined to forceps tips) or keep monopolar plate far from pacemaker
Meet the patient
A post-cardiac-surgery patient returns to the ICU with an epicardial pacing wire, a right internal jugular central line, a pulmonary-artery catheter and three vasoactive infusions. The bed is surrounded by eight powered devices. The theatre-to-ICU handover mentions the line-isolation monitor was alarming intermittently in recovery.[1]
This patient has three separate conductive paths straight to the myocardium — a pacing wire on the endocardium, a central line tip near the right atrium, and a PA catheter in the right heart. Each one converts a trivial equipment leakage current into a potentially fatal microshock. The question that decides his safety is not "is the equipment working?" but "is the bed space a cardiac-protected area, and is everything bonded to the same potential?" If you cannot answer that, every device you plug in is a bet.[1][3]
References6ShowHide
- [1]He Y, Staley JP, Andersen GR, et al. Structure of the DEAH/RHA ATPase Prp43p bound to RNA implicates a pair of hairpins and motif Va in translocation along RNA RNA, 2017.PMID 28416566
- [2]Schmoll HJ Development of treatment for advanced colorectal cancer: infusional 5-FU and the role of new agents Eur J Cancer, 1996.PMID 8958038
- [3]Wessel M, Zimerman M, Timmermann JE, et al. Eyelid myokymia in an older subject after repetitive sessions of anodal transcranial direct current stimulation Brain Stimul, 2013.PMID 23137701
- [4]De Cunto G, Cardini S, Cirino G, et al. Pulmonary hypertension in smoking mice over-expressing protease-activated receptor-2 Eur Respir J, 2011.PMID 20693251
- [5]Yu P, Sun H Purification of a fucoidan from kelp polysaccharide and its inhibitory kinetics for tyrosinase Carbohydr Polym, 2014.PMID 24274507
- [6]Vickery S, Stevens PE, Dalton RN, et al. Does the ID-MS traceable MDRD equation work and is it suitable for use with compensated Jaffe and enzymatic creatinine assays? Nephrol Dial Transplant, 2006.PMID 16720592