Pressure and temperature measurement

Also known as Pressure measurement · Pressure transducer · Arterial line · Damping · Fast-flush test · CVP waveform · Pulmonary artery catheter · Temperature measurement · Thermistor · Thermocouple · RTD

Blood pressure and body temperature are the two most frequently measured physiological variables in anaesthesia, and the physics of how they are measured determines whether the numbers on the screen are trustworthy. The framework rests on six exam-critical ideas. First, PRESSURE is force per unit area, with a defined set of clinical units and conversions; it is measured by mechanical devices (the Bourdon gauge for high-pressure cylinders, the aneroid gauge and the liquid manometer for anaesthetic-machine and venous pressures) and by electronic transducers for invasive blood pressure. Second, invasive blood pressure is measured by a PRESSURE TRANSDUCER — a device that converts the mechanical pressure of a fluid column into an electrical signal, typically a piezoresistive strain gauge bonded to a diaphragm whose resistance changes with deformation, detected by a Wheatstone bridge. Third, the transducer must be ZEROED to atmospheric pressure and LEVELLED to the mid-axillary line at the right atrium, because each centimetre of height difference introduces a hydrostatic pressure error of about 0.75 to 1 mmHg. Fourth, the dynamic response of the catheter-tubing-transducer system depends on its NATURAL FREQUENCY and DAMPING, assessed by the FAST-FLUSH test: under-damping overshoots (spurious high systolic), over-damping lags (under-read systolic, over-read diastolic), and mean pressure is preserved in both. Fifth, temperature is measured by several physical principles — a THERMISTOR (resistance falls as temperature rises), a THERMOCOUPLE (Seebeck voltage from two dissimilar metals), a platinum RTD (resistance rises with temperature, the laboratory reference), and an INFRARED detector — and the choice of monitoring SITE matters because core temperature is best estimated from oesophageal, nasopharyngeal, pulmonary-artery, tympanic or bladder sites while skin lags and under-reads. Sixth, general anaesthesia abolishes thermoregulation and produces the three phases of perioperative heat loss — redistribution (the steep first-hour fall), linear decline, and plateau — and prevention of inadvertent hypothermia by prewarming and forced-air warming is a quality indicator. Built on the dynamic-response requirements paper (Gardner 1981), the fast-flush test validation (Kleinman 1992), the arterial catheter complications review (Scheer 2002), the AHA blood-pressure-measurement statement (Pickering 2005), the transradial-access ultrasound guidance (Hamilton 2024), the PAC-Man trial (Harvey 2005), the perioperative thermoregulation reviews (Sessler 2008, 2016), the peripheral-thermometer accuracy meta-analysis (Niven 2015), the continuous-temperature-methods comparison (Ehlers 2025), and the four landmark hypothermia-outcomes trials (Frank 1997, Kurz 1996, Schmied 1996, Heier 1991).

high16 referencesUpdated 26 July 202622 min readVerification in progress

Practise this topic

  • 8 MCQs

Your progress

Saved on this device.

Practise this topic8 MCQs with explanations

Target exams

ANZCAFRCAABAEDAICFCAIFCA_SA

Red flags

  • Pressure is force per unit area. Unit conversions: 1 kPa = 7.5 mmHg = 10.2 cmH2O; 1 mmHg = 1.36 cmH2O = 0.133 kPa; 1 atm = 101.325 kPa = 760 mmHg = 1.013 bar.
  • Invasive BP uses a STRAIN-GAUGE (piezoresistive) pressure transducer: a diaphragm deforms, changing the resistance of bonded strain gauges, detected by a Wheatstone bridge, converting hydraulic pressure to an electrical signal.
  • The transducer must be ZEROED to atmospheric pressure and LEVELLED to the mid-axillary line (right atrium). Each cm of height error introduces about 0.75 to 1 mmHg of hydrostatic pressure error — a SYSTEMATIC (bias) error, not random.
  • DAMPING: under-damped = overshoot (spurious high systolic, low diastolic); over-damped = lag (under-read systolic, over-read diastolic). MEAN pressure is preserved in both. The fast-flush (square-wave) test assesses the dynamic response.
  • CVP waveform: three waves (a, c, v) and two descents (x, y). a = atrial contraction (after P wave); c = tricuspid bulging (start of QRS); v = atrial filling (after T wave). Cannon a waves = complete heart block; large v waves = tricuspid regurgitation; absent a waves = atrial fibrillation.
  • THERMISTOR = resistance FALLS as temperature rises (negative coefficient). THERMOCOUPLE = two dissimilar metals, Seebeck voltage proportional to temperature difference. RTD (platinum) = resistance RISES with temperature (positive coefficient), the most accurate. IR = non-contact.
  • Core temperature sites: oesophageal lower third, nasopharyngeal, pulmonary artery, tympanic (contact), bladder. Skin lags and under-reads by 1 to 2 degrees. The three phases of perioperative heat loss: redistribution (1 to 1.5 degrees, first hour, NOT heat loss), linear decline (0.5 to 1 degree/hr), plateau.
On this page
Study tools

Your progress

Saved on this device.

Practise this topic8 MCQs with explanations

Target exams

ANZCAFRCAABAEDAICFCAIFCA_SA

Red flags

  • Pressure is force per unit area. Unit conversions: 1 kPa = 7.5 mmHg = 10.2 cmH2O; 1 mmHg = 1.36 cmH2O = 0.133 kPa; 1 atm = 101.325 kPa = 760 mmHg = 1.013 bar.
  • Invasive BP uses a STRAIN-GAUGE (piezoresistive) pressure transducer: a diaphragm deforms, changing the resistance of bonded strain gauges, detected by a Wheatstone bridge, converting hydraulic pressure to an electrical signal.
  • The transducer must be ZEROED to atmospheric pressure and LEVELLED to the mid-axillary line (right atrium). Each cm of height error introduces about 0.75 to 1 mmHg of hydrostatic pressure error — a SYSTEMATIC (bias) error, not random.
  • DAMPING: under-damped = overshoot (spurious high systolic, low diastolic); over-damped = lag (under-read systolic, over-read diastolic). MEAN pressure is preserved in both. The fast-flush (square-wave) test assesses the dynamic response.
  • CVP waveform: three waves (a, c, v) and two descents (x, y). a = atrial contraction (after P wave); c = tricuspid bulging (start of QRS); v = atrial filling (after T wave). Cannon a waves = complete heart block; large v waves = tricuspid regurgitation; absent a waves = atrial fibrillation.
  • THERMISTOR = resistance FALLS as temperature rises (negative coefficient). THERMOCOUPLE = two dissimilar metals, Seebeck voltage proportional to temperature difference. RTD (platinum) = resistance RISES with temperature (positive coefficient), the most accurate. IR = non-contact.
  • Core temperature sites: oesophageal lower third, nasopharyngeal, pulmonary artery, tympanic (contact), bladder. Skin lags and under-reads by 1 to 2 degrees. The three phases of perioperative heat loss: redistribution (1 to 1.5 degrees, first hour, NOT heat loss), linear decline (0.5 to 1 degree/hr), plateau.
The one-line answer

Pressure is force per unit area, measured by the Bourdon gauge for high-pressure cylinders, the aneroid gauge for machine supply pressures, and the piezoresistive strain-gauge transducer for invasive lines — a diaphragm, four strain gauges, and a Wheatstone bridge. Zero to atmosphere, level to the mid-axillary line, and trust the mean when damping distorts the trace. Temperature rests on four sensing principles: the thermistor (resistance falls), the thermocouple (Seebeck voltage), the RTD (resistance rises, the most accurate), and infrared (non-contact, fails in shock).[1][9]

Meet the patient

A 68-year-old man is in theatre for an emergency laparotomy. The arterial line reads 240 over 60 and the registrar reaches for a beta-blocker. The fast-flush test rings like a struck bell — many oscillations, spurious systolic. The true mean is 90 mmHg, the one value that survives every damping state.[1]

Two questions decide every monitoring decision you will make: is this number trustworthy? — the fast-flush test answers that for pressure, the sensing principle and site answer it for temperature — and am I measuring what I think I am? Hold those two questions and the rest of the topic slots into place.[1]

Anaesthesia Pro

Continue reading

You have read the opening of this topic. The complete unit — every section and its primary-source references — is part of the Anaesthesia fellowship library.

Sign in to continueSee pricing
References16ShowHide
  1. [1]Gardner RM Direct blood pressure measurement--dynamic response requirements Anesthesiology, 1981.PMID 7469106
  2. [2]Kleinman B, Powell S, Kumar P The fast flush test measures the dynamic response of the entire blood pressure monitoring system Anesthesiology, 1992.PMID 1466471
  3. [3]Scheer B, Perel A, Pfeiffer UJ Clinical review: complications and risk factors of peripheral arterial catheters used for haemodynamic monitoring in anaesthesia and intensive care medicine Crit Care, 2002.PMID 12133178
  4. [4]Pickering TG, Hall JE, Appel LJ, et al Recommendations for blood pressure measurement in humans: an AHA scientific statement from the Council on High Blood Pressure Research Professional and Public Education Subcommittee J Clin Hypertens (Greenwich), 2005.PMID 15722655
  5. [5]Hamilton GW Ultrasound Guidance for Transradial Access in the Cardiac Catheterisation Laboratory: A Systematic Review of the Literature and Meta-Analysis Heart Lung Circ, 2024.PMID 38871531
  6. [6]Harvey S, Harrison DA, Singer M, et al Assessment of the clinical effectiveness of pulmonary artery catheters in management of patients in intensive care (PAC-Man): a randomised controlled trial Lancet, 2005.PMID 16084255
  7. [7]Harvey SE, Welch CA, Harrison DA, et al Post hoc insights from PAC-Man--the U.K. pulmonary artery catheter trial Crit Care Med, 2008.PMID 18496358
  8. [8]Sessler DI Perioperative thermoregulation and heat balance Lancet, 2016.PMID 26775126
  9. [9]Sessler DI Temperature monitoring and perioperative thermoregulation Anesthesiology, 2008.PMID 18648241
  10. [10]Insler SR, Sessler DI Perioperative thermoregulation and temperature monitoring Anesthesiol Clin, 2006.PMID 17342966
  11. [11]Niven DJ, Gaudet JE, Laupland KB, et al Accuracy of peripheral thermometers for estimating temperature: a systematic review and meta-analysis Ann Intern Med, 2015.PMID 26571241
  12. [12]Ehlers UE, Ulmer J, Keller M, et al Comparison of continuous temperature measurement methods in the intensive care unit: standard bladder catheter measurements versus non-invasive transcutaneous sensors J Clin Monit Comput, 2025.PMID 39066870
  13. [13]Frank SM, Fleisher LA, Breslow MJ, et al Perioperative maintenance of normothermia reduces the incidence of morbid cardiac events. A randomized clinical trial JAMA, 1997.PMID 9087467
  14. [14]Kurz A, Sessler DI, Lenhardt R Perioperative normothermia to reduce the incidence of surgical-wound infection and shorten hospitalization. Study of Wound Infection and Temperature Group N Engl J Med, 1996.PMID 8606715
  15. [15]Schmied H, Kurz A, Sessler DI, et al Mild hypothermia increases blood loss and transfusion requirements during total hip arthroplasty Lancet, 1996.PMID 8569362
  16. [16]Heier T, Caldwell JE, Sessler DI, et al Mild intraoperative hypothermia increases duration of action and spontaneous recovery of vecuronium blockade during nitrous oxide-isoflurane anesthesia in humans Anesthesiology, 1991.PMID 1673591

Test yourself

Practise what you just read

  • 8 MCQs
Anaesthesia Pro