Electricity powers every device in the operating theatre and carries the two specific hazards of macroshock (whole-body shock from mains contact) and microshock (a tiny current delivered directly to the heart through a catheter causing ventricular fibrillation at currents as low as 100 microamperes). The framework rests on six exam-critical ideas. First, the four fundamental quantities are CHARGE (the coulomb, C), CURRENT (the ampere, A, equal to one coulomb per second), VOLTAGE (the volt, V, the electrical potential difference or electromotive force) and RESISTANCE (the ohm); they are related by OHM'S LAW (V equals I times R) and by the power equation (wattage equals volts times amps). Second, mains electricity is ALTERNATING CURRENT (AC), cycling at 50 hertz in Australasia and the UK (60 hertz in North America); AC is more dangerous than direct current (DC) at the same voltage because the alternating cycle can induce tetanic muscle contraction (preventing release from a live wire) and because it crosses zero twice per cycle, making the heart vulnerable to ventricular fibrillation at the crossover. Third, the body's electrical safety depends on EARTHING (a low-resistance path to ground that diverts fault current away from the patient and trips the fuse), FUSES and CIRCUIT BREAKERS (which break the circuit when current exceeds a set limit), RESIDUAL CURRENT DEVICES (an RCD detects any difference between the current flowing out and returning, tripping in milliseconds at about 5 to 30 milliamps of leakage), and ISOLATED (FLOATING) CIRCUITS (the mains supply is separated from earth by a transformer so that a single fault cannot complete a circuit through the patient to ground). Fourth, MACROSHOCK is the whole-body shock from contact with a live conductor through the skin; the severity depends on the current (1 milliampere is the threshold of perception, 10 to 20 milliamps is the let-go threshold with muscle tetany, 50 to 100 milliamps causes respiratory failure, and over about 100 milliamps causes ventricular fibrillation). Fifth, MICROSHOCK is a very small current (as low as 100 microamperes) delivered DIRECTLY to the myocardium through a low-resistance pathway such as a saline-filled central venous catheter, a pulmonary artery catheter or a pacemaker wire, causing ventricular fibrillation at a current far below the macroshock threshold; this is why all patient-connected equipment must be earthed, isolated and leakage-tested, and why a saline-filled central line is never connected to non-isolated equipment. Sixth, DIATHERMY (electrosurgery) uses high-frequency AC (above about 500 kilohertz) to cut and coagulate tissue without neuromuscular stimulation; MONOPOLAR diathermy passes current from an active electrode through the patient to a large dispersive plate (return electrode), while BIPOLAR current passes between the two jaws of a forceps and does not require a return plate; diathermy burns, surgical fires (the electrosurgical spark igniting an oxygen-enriched atmosphere with a fuel source) and electromagnetic interference with pacemakers and ICDs are the anaesthetic hazards. Built on the electrical-safety review (Burgess 2019), the electrosurgical-hazards review (Pearce 1985), the electrosurgery-principles review (Vilos 2013), the surgical-smoke review (Larivee 2026), the electrical-injury-reconstruction study (Zhang 2026), the electrical-injury-cataract study (Wang 2026), the electrical-burn-amputation study (Sankhla 2026), the device-endocarditis report (Ohev Shalom 2026), the operating-room-fire review (Mortada 2024), and the skin-prep-fire study (Jones 2017).
Practise this topic
On this page
Study tools
Your progress
Saved on this device.
Target exams
Red flags
- OHM'S LAW: V equals I times R (voltage equals current times resistance). Power: P equals V times I equals I squared times R. These two equations govern every electrical device in theatre.
- MACROSHOCK thresholds: 1 mA perception, 10-20 mA let-go (muscle tetany), 50-100 mA respiratory failure, greater than 100 mA ventricular fibrillation. AC is more dangerous than DC at the same voltage.
- MICROSHOCK: a current as low as 100 MICROAMPERES (0.1 mA) can cause VF when delivered DIRECTLY to the myocardium through a central line, PA catheter or pacemaker wire. This is 1000 times lower than the macroshock VF threshold.
- Electrical safety: earthing (diverts fault current to ground, trips fuse), RCD (trips at 5-30 mA leakage in milliseconds), ISOLATED/FLOATING circuits (a single fault cannot complete a circuit through the patient).
- DIATHERMY: monopolar (active electrode through patient to return plate) vs bipolar (between forceps jaws, no return plate). Radiofrequency above 500,000 Hz heats without neuromuscular stimulation.
- SURGICAL FIRE TRIANGLE: an ignition source (diathermy spark), an oxidiser (oxygen or nitrous oxide enriched atmosphere) and a fuel (skin prep alcohol, drapes, hair). Avoid pooling alcohol preps around the diathermy site.
Meet the patient
A patient in ICU has a central venous catheter and a pulmonary artery catheter in situ. A nurse touches the monitor and the catheter simultaneously, and a tiny leakage current — imperceptible to her skin — passes through the catheter directly to the myocardium. The patient goes into ventricular fibrillation. No one felt a thing. That is microshock: a direct electrical connection to the patient puts them at increased shock risk, and additional, more direct electrical pathways to the heart make them even more vulnerable, especially during monitoring in the operating room or intensive care unit.[1]
The two questions that decide electrical danger are: how much current is flowing? (Ohm's law: I equals V divided by R) and where is the current going? (the path — through skin and muscle, or straight to the heart). Voltage slogans are meaningless without both.[1]
References10ShowHide
- [1]Burgess RC Electrical safety. Handb Clin Neurol, 2019.PMID 31277877
- [2]Pearce J Current electrosurgical practice: hazards. J Med Eng Technol, 1985.PMID 4009682
- [3]Vilos GA, Rajakumar C Electrosurgical generators and monopolar and bipolar electrosurgery. J Minim Invasive Gynecol, 2013.PMID 23659748
- [4]Larivee J, et al. Surgical Smoke Evacuation: An Integrative Review With Implications for Practice and Sustainment in the Military Health System. Mil Med, 2026.PMID 42463130
- [5]Zhang Y, et al. Secondary reconstruction of tendons and nerves improves long-term function in electrical burns of the wrist: A retrospective cohort study Burns, 2026.PMID 42296601
- [6]Wang Y, et al. Cataract after electrical injury: clinical characteristics and surgical management outcomes, a case report of three cases BMC Ophthalmol, 2026.PMID 42288764
- [7]Sankhla A, et al. Functional Outcomes Following Amputation in Electrical Burn Injuries: A Prospective Observational Study Utilizing the Sickness Impact Profile (SIP) and International Classification of Impairments, Disabilities, and Handicaps (ICIDH) Frameworks Cureus, 2026.PMID 42255776
- [8]Ohev Shalom R, et al. Tricuspid valve infective endocarditis due to ESBL producing E.coli associated with AV block requiring epicardial pacing: A case report J Cardiothorac Surg, 2026.PMID 42363188
- [9]Mortada H, et al. Preventing and Managing Operating Room Fires in Plastic Surgery: A Review of Incidence, Risk Factors, and Recommendations With Case Experiences. J Burn Care Res, 2024.PMID 38158891
- [10]Jones EL, et al. Operating Room Fires and Surgical Skin Preparation. J Am Coll Surg, 2017.PMID 28189662